index previous next

Signal and Photon SNR of Atomic Emission Spectrometry


Click to see a larger view.

[Cell definitions and equations]

Simplified model of trace metal analysis in solution by a flame or plasma emission spectrometry. Includes the effects of solution transport, nebbulization, excitation, collection of a fraction of the resulting light, detection with a monochromator and photomultiplier tube, and computation of the photon signal-to-noise ratio. The main purpose of this simulation is to demonstrate that atomic emission spectroscopy is capable of trace analysis of solutions under the right conditions and to illustrate how the signal-to-noise ratio varies with temperature and with the excitation energy of the element.

This model can not be expected to predict signals and signal-to-noise ratios accurately because of its many simplifying assumptions: thermal equilibrium is assumed; overall atomization efficiency includes nebulization efficiency and free-atom fraction (both assumed independent of temperature); no ionization or compound formation; self-absorption is ignored; only photon noise considered, no background emission is assumed. However, order-of-magnitude predictions may be obtained in many cases.

Note: You may adjust the temperature and the wavelength with either the sliders or by typing into the inputs column.

Download links: AES.wkz; AES.hqx
Having trouble getting this to work? Download the complete system of modules for PC or Mac.
[Return to Index]


Cell definitions and equations:

Inputs:		
Concentration, cg	µg/mL
Solution flow rate, F	mL/sec
Overall atomization efficiency, epsilon	
Total gas flow rate, Q	L/sec
Flame/plasma temperature, T	K
Relative # moles burnt gases, nT	
Relative # moles unburnt gases, nRT		
Formula weight of analyte, MW	g/mole
Wavelength of line, lambda	nm
Einstein A coefficient, Aji	sec-1
Statistical weight of lower state,glower	
Statistical weight of upper state, gupper		
Path length, l		cm
Quantum efficiency of photocathode, Klambda	
Photomultiplier gain, m	
Slit width, W	cm
Slit height, H	cm
Solid angle of monochromator, omega	sr
Monochromator transmission factor, Top	

Outputs:		
analyte molarity, c	=0.001*cg/MW	
frequency of transition, fo	=(2.998E+17)/lambda	
energy of transition, E	=(6.6261E-34)*freq	
Boltzman factor	=exp(-E/(T*1.3805E-23))	
gas expansion factor, ef	=(nT*T)/(nRT*298)	
Number in upper state , nupper	=nlower*(gupper/glower)*exp(-E/(T*1.3805E-23)) 	
Number in lower state, nlower	=6.00E+17*F*epsilon*c/(Q*ef)	
Emission radiance, Be	=Aji*E*nupper*l/(4*pi())	
radiant cathode sensitivity, Rlambda	=(Klambda*1.602E-19)/E	
photoanodic current, Ie	=m*Rlambda*W*H*omega*Top*Be	
photon flux on detector, PhotFlux	=Aji*nupper*W*H*omega*Top/(4*pi())	
photoelectron emission rate, Rcp	=Klambda*PhotFlux	
Signal-to-photon-noise ratio, SNR	=Rcp/sqrt(Rcp)	

(c) 1994, 2000, Prof. Tom O'Haver , Professor Emeritus, The University of Maryland at College Park. Comments, suggestions and questions should be directed to Prof. O'Haver at to2@umail.umd.edu.